Hantek DSO 4202C display the wrong tension

[Copy link]

1

Threads

1

Posts

1

Credits

新手上路

Rank: 1

Credits
1
Jump to specified page
1#
Post time 2017-8-31 06:22:42 | Show the author posts only Reply Awards |Descending |Read mode
Hello,

It seems that I have some issue with my DSO 4202C shortly after purchase.
After using the DSO 4202 C for the first time outside initial function testing including the probe calibration, the amplitude of the voltage was just completely wrong; 56V 40Hz square instead of 5V 40Hz square (the 5V was checked with another device). When connecting the probe to the 5V 1kHZ reference signal, the amplitude is still 56V peak, frequency is right. It seems that the tension is locked to 56V, even if I change it with a signal generator.

I run the calibration again, but no improvement.
I tried the firmware upgrade, but the version on my device is more recent than the one available on the website.

What should I do, is it a bug or a faulty device? It is quite annoying after just 13 boots of the device.

Technical info:
SW - 3.50.0(170628.0)
HW - 1020xffffa000
serial T1G/0490011643
only 13 boot, uncluding the tests above.

Regards,

Reply

Use magic Report

24

Threads

119

Posts

119

Credits

注册会员

Rank: 2

Credits
119
2#
Post time 2017-8-31 06:34:05 | Show the author posts only
Edited by Oleg10011001 at 2017-8-31 06:39

awuersch, most likely you just do not know how to use your oscilloscope and could not properly configure it. RTFM!!! I think, in the oscilloscope settings, it is not correctly indicated which divider is on the probe. Most likely on the probe, the divider switch is in the x1 mode, and in the oscilloscope settings it is indicated that the divider x10 is used. You need to either change the oscilloscope settings, or switch the switch on the probe to x10 mode.




167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
3#
Post time 2017-9-1 11:28:20 | Show the author posts only



Sir, please change the probe attenuation factor to 1× on the scope as the picture.



Make sure the probe attenuation factors on the scope and the proble are the same.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-5-2 20:54 GMT+8 , Processed in 0.184080 second(s), 21 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list