Hantek DSO5202BM: Confusing vertical shift waveform.

[Copy link]

24

Threads

119

Posts

119

Credits

注册会员

Rank: 2

Credits
119
Jump to specified page
1#
Post time 2017-1-10 14:23:00 | Show the author posts only Reply Awards |Descending |Read mode
There is a strange shift in the vertical waveforms obtained from the second channel when the oscilloscope is set at more than 2V per cell. For example, look at the waveform in the following screenshots. In all of these cases, both the probe was connected to a generator, oscilloscope calibration, and both channels markers moved strictly in the center of the screen by pressing the corresponding "Vertical position".









Reply

Use magic Report

167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
2#
Post time 2017-1-17 15:22:37 | Show the author posts only
Please disconnect any probes or cables to the channels and press the UTILITY button, select Self Calibration option and follow the directions on the screen.

24

Threads

119

Posts

119

Credits

注册会员

Rank: 2

Credits
119
3#
 Author| Post time 2017-1-17 22:13:54 | Show the author posts only
Thanks, this helped!