41#
 Author| Post time 2016-9-19 08:44:33 | Show the author posts only
Edited by WisdomAugust at 2016-9-19 08:45

3.6 To change the Volts/Div control sensitivity

When you need to adjust the amplitude/div setting in smaller increments, you can change the sensitivity of the vertical scale control.
?If the channel’s menu is not currently displayed, press the channel key (CH1 MENU, or CH2 MENU).
?In the Channel menu, press Volts/Div to toggle between:

?Coarse The vertical scale knob changes the amplitude/div setting in a 1- 2- 5 step sequence from 2 mV/div to 10 V/div (with “1X” probe attenuation).
?Fine Also known as vernier, the vertical scale knob changes the amplitude/div setting in small steps between the normal (coarse scale) settings.


58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
42#
 Author| Post time 2016-9-20 08:48:06 | Show the author posts only
3.7 To specify the probe attenuation
For correct measurements, you must match the oscilloscope’s probe attenuation factor settings with the attenuation factors of the probes being used.
The probe attenuation factor setting changes the vertical scaling of the oscilloscope so that the measurement results reflect the actual voltage levels at the probe tip.
? If the channel’s menu is not currently displayed, press the channel key ( CH1 MENU, or CH2 MENU).
? In the Channel menu, press Probe.
? Continue pressing the Probe softkey to select between:
l 1X For 1:1 probes.
l 10X For 10:1 probes.
l 100X For 100:1 probes.
l 1000X For 1000:1 probes

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
43#
 Author| Post time 2016-10-8 11:56:09 | Show the author posts only



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
44#
 Author| Post time 2016-10-9 08:31:23 | Show the author posts only
3.9 Using Math Function Waveforms
3.9.1 The math functions control allows the selection of the math functions:
? Add.
? Subtract.
? Multiply.
? Divide.
? FFT (Fast Fourier Transform).
The mathematical result can be measured using the grid and cursor controls.
The amplitude of the math waveform can be adjusted using a menu item selection in the Math menu.
The math scale setting is shown at the bottom of the Math menu.

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
45#
 Author| Post time 2016-10-10 09:35:45 | Show the author posts only
Edited by WisdomAugust at 2016-10-10 09:38

3.9.2 To add, subtract, multiply, or divide waveforms
1. Press MATH MENU.
2. In the Math menu, press Operation.
3. Continue pressing the Operation softkey to select “CH1+ CH2”, “CH1- CH2”, “CH2-CH1”, “CH1 x CH2”, “CH1/CH2”, or “CH2/CH1”.
4. Press Position softkey and turn multifuction knob to adjust the position of the math waveform.
5. Press Scale, and turn multifuction knob to adjust the math scale.


58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
46#
 Author| Post time 2016-10-19 08:46:52 | Show the author posts only
3.9.3 To display the frequency domain using FFT

The FFT math function mathematically converts a time- domain waveform
into its frequency components. FFT waveforms are useful for finding the
harmonic content and distortion in systems, for characterizing noise in DC
power supplies, for analyzing vibration, for estimating the impulse responses in systems, and for measuring the harmonic in power cords.
To display a waveform’s FFT:
1.Press MATH MENU.
2.In the Math menu, press Operation.
3.To select “FFT”.
4. Press Sourcesoftkey to select the desired input channel.

NOTE
? The FFT of a waveform that has a DC component or offset can cause incorrect FFT waveform magnitude values. To minimize the DC component, choose AC Coupling on the source waveform.
? To reduce random noise and aliasing components (in repetitive or single-shot waveforms), set the oscilloscope acquisition mode to averaging.


58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
47#
 Author| Post time 2016-10-20 08:37:20 | Show the author posts only

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
48#
 Author| Post time 2016-10-20 08:37:20 | Show the author posts only

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
49#
 Author| Post time 2016-10-21 10:27:25 | Show the author posts only
If the number of cycles is non-integral, the YT waveform starts and ends at different amplitudes and transitions between the start and end points will cause discontinuities in the signal that introduces high-frequency transients.

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
50#
 Author| Post time 2016-10-24 08:48:50 | Show the author posts only
Edited by WisdomAugust at 2016-10-24 08:50

Applying a window to the YT waveform changes the waveform so that the start and stop values are close to each other, which reduces the discontinuities.




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-5-5 15:27 GMT+8 , Processed in 0.166946 second(s), 19 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list