Another passive probe is a low-resistance passive probe,
which is the least commonly used probe.
The biggest advantage of the low-resistance passive probe
is that it has relatively high test bandwidth (1G-6GHz) with
the price of an ordinary high-impedance passive probe.
The main drawback is that the input impedance is low
(only 500 ohm or 1k ohm), if connect to the circuit, it may
still have a little impact on the measured signal.
The following is its duagrammatic layout.




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
26#
 Author| Post time 2017-6-22 13:17:02 | Show the author posts only
There are two special types of high impedance passive probes.
One is a high-voltage probe, the attenuation ratio up to 100:
1 or 1000: 1, so the measurement voltage range is large;
another is a type of 1: 1 probe, i.e., there’s no attenuation
of the signal when input the oscilloscope. Not like 10: 1
probe which need the oscilloscope to enlarge the display,
so the noise of the oscilloscope itself is not amplified,
and the measured of noise will be much smaller, which use
a lot in the small signal and power ripple measurement.

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
25#
 Author| Post time 2017-6-21 08:11:37 | Show the author posts only
Edited by WisdomAugust at 2017-6-22 13:22

The simpler probe needs to set the oscilloscope's attenuation manually
to get the correct display. More probes have an auto-detect pin at the
end which connect to the oscilloscope. It can read the attenuation
ratio of the probe through the pin when the probe is plugged in
the oscillsocope, and adjust the display ratio automatically.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
24#
 Author| Post time 2017-6-20 09:21:08 | Show the author posts only
Rprobe improves the frequency response and produces a
divider with oscilloscope input resistance. The so-called
10: 1 divider is the actual measured voltage of the
oscilloscope is 1/10 of the voltage at the front end
of the probe, i.e. the signal through the probe will
have a 10 times attenuation.

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
23#
 Author| Post time 2017-6-19 08:24:06 | Show the author posts only
Edited by WisdomAugust at 2017-6-22 13:21

When the probe is connected to different channel of
the oscilloscope, it can compensate Cscope changes
by adjusting the Ccomp. Almost all oscilloscopes
provide a low frequency square wave output,
which can connect to the probe to adjust
the shape of the saqure wave to adjust
the Cscope.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
22#
 Author| Post time 2017-6-16 08:10:00 | Show the author posts only
A condition for the probe to produce a flat gain in
the band is to meet Rprobe * Cprobe = Rscope * Cscope.
Cscope is the parasitic capacitance of the oscilloscope,
so it can only be controlled in a certain range,but
can not be precisely controlled, i.e. The Cscope value
of different oscilloscopes or different channels on
oscilloscope will be different. In order to compensate
for changes in different channels Cscope, at the end of
the probe connected to the oscilloscope, where will be
at least an adjustable capacitor-Ccomp.

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
21#
 Author| Post time 2017-6-15 08:27:28 | Show the author posts only
To facilitate the measurement, the probe usually
has around 1 meter length. If not adding the matching
circuit, it is hard to imagine that the probe can
provide hundreds of megabytes of bandwidth. The input
parasitic capacitance of the oscilloscope affects the
bandwidth as well.

To improve the high frequency corresponding of the probe,
there will be a corresponding matching circuit on the
probe’s front-end. The most typical is a Rprobe and
Cprobe parallel structure.

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
20#
 Author| Post time 2017-6-12 08:27:30 | Show the author posts only
Edited by WisdomAugust at 2017-6-22 13:21

Sorted by  the size of the input impedance is divided into
two types: high impedance passive probe and low resistance
passive probe.

High-impedance passive probe is the one that we usually call
passive probe, which is the most widely used probe, basically
every person using the oscilloscope have used this type of probe.
When the high-impedance passive probe is connected to the
oscilloscope, the input impedance of the oscilloscope is
required to be 1M ohms. The following is a schematic diagram
of a 10: 1 high impedance passive probe.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
19#
 Author| Post time 2017-6-10 10:33:53 | Show the author posts only
Edited by WisdomAugust at 2017-6-22 13:20

According to the power supply oscilloscope probe can be
divided into passive probe and active probe, sorted by
the measured signal type the probe can be divided into
voltage probe, current probe, optical probe etc. The
so-called passive probe indicates the entire probe by
composed by passive components, including resistors,
capacitors, cables etc; well the active probe inside
has general amplifier, and the amplifier need to be
powered, so named active probe.

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1277

Posts

1357

Credits

金牌会员

Rank: 6Rank: 6

Credits
1357
18#
 Author| Post time 2017-6-7 08:16:32 | Show the author posts only
Broadly speaking, the test leads belongs to probe as well,
like BNC or SMA cable. This type of probe is cheap and high
performance. But the use of test leads connection require the
to be tested circuit has BNC or SMA interface, thus their application
is limited, mainly for radio frequency and microwave signal testing.
For digital or general signal testing, should use a special probe.