CC-65 AC/DC CURRENT CLAMP OPERATOR’S MANUAL

[Copy link]

61

Threads

104

Posts

345

Credits

中级会员

Rank: 3Rank: 3

Credits
345
Post time 2016-3-8 09:11:34 | Show all posts |Read mode
CC-65AC/DC CURRENT CLAMP
OPERATOR’SMANUAL
INTRODUCTION
When measuring the AC/DC Current with the Current Clamp, a multimeteror scope can directly measure the current up to 65 A , with a frequency up to20KHz. In this method, there is no need to break a circuit or to affect theisolation to measure the current.
The Current Clamp is not only available for the detection or monitor ofthe possibly existing creepage, but for the measurement with little gap usingits extended jaws. However, it is necessary to turn the zero adjustment knobson the clamp until the multimeter or scope reads zero.
APPLICATIONPROCEDURE
1. When connected with anymulti-meter whose input impedance is 10k ohms minimum,insert the black bananaplug into the COM jack and the red banana plug into the V-Ω jack.When connectedwith a scope,connect the BNC connector of the clamp to a scope BNC connector.
2. Set the power switch form“Off” to “ON” position, that is to set the multimeter range to 1mV/10mAor 1mV/100mA. Then, the green LED willbe lighting, indicating that the clamp is switched on..
3. when the current is notgreater than 2 ampere ,please set the current clamp range to 1mV/10mA and setthe multi-meter range to 200mV AC or 200 mV DC, which depends on the type ofthe current.
4. Before measuring the DCcurrent, it is needed to turn the zero adjustment knobs on the clamp until themultimeter reads zero. That means pushing down the zero adjustment knobs on theclamp until the multimeter reads zero. Then let go hold of the zero adjustmentknobs.When using a scope,set DC coupled mode to measure DC current.
5. Clamp the jaws around thecurrent-carrying conductor and operate as above.
6. When the current clamprange is set to 1mV/10mA, the actual currentis ten times of the measured current whose unit is mA. For example, themeasured current is 10 mV and The actual current of the current-carryingconductor is 10×10=100mA.
7. When the current clamprange is set to 1mV/100mA, the actual currentis one hundred times of the measured current whose unit is mA. For example, themeasured current is 5 mV and The actual current of the current-carryingconductor is 5×100=500mA.
APPLICATOIN NOTES
1. In the case of DC current, the output is positive when the currentflows from the upside to the underside of the clamp. The red banana plug ispositive.
2. In the case of DC current measurement, the current clamp can notpossibly zero properly, due to the hysteresis effect. To eliminate this, openand close the jaws several times and then zero again.
APPLICATIONSAFETY
1. The voltage of the conductor measured with the current clamp shouldbe little than 300V DC or 240V rms AC,
2. We are forbidden to measure the conductor whose insulation has breakdown or wear down, in order to avoid doing harm to us.
SPECIFICATIONSGENERAL
Captured Conductor diameter Size: 9mm maximum
Low Battery Indicator: Red LED lighting
Operating Temperature: 0℃ to 50℃, 70% R.H.
Storage Temperature: -20℃~+70℃, 80% R.H.
Battery type: 9V DC, NEDA 1604, 6F22, 006P
Battery Life: 100 hours typical withcarbon-zinc
Weight: 250 gram typical
Dimensions: 195 mm (H) x 70 mm (W) x 33 mm(D)
Output: wire with a straight banana plug orBNC connector
ELECTRICAL (At23±5℃, 70% R.H. maximum)
Efficient Measure Range:
1mV/10mA: 10mA to 2 A, multimeter range is set to DC or rms AC 200mV
1mV/100mA: 100mA to 20 A, multimeter range is set to DC or rms AC 200mV
                            20A to65 A, multimeter range is set to DC or rms AC 2V
Precision:
System Accuracy: The Precision of the current clamp add the Precisionof the digital multimeter or scope.
For example, if the conductor carries 100 mA current, the current clamprange will be set to 1mV/10mA and the output is 10mV. Assuming that thePrecision is 1.5%, the output is10.15mV maximum and 9.85mV minimum.
If the digital multimeter connected with current clamp is set to 200mVand the Precision is 0.5%,,the readings is10.20mV maximum and 9.8mV minimum.(10.15mV×(1+0.5%)=10.20mV9.85mV×(1-0.5%)=9.80mV)
Accuracy:
         DC Current: 1m/10mA
±(1.5%±5mA) 10mA ~ 20A
         DC Current: 1mV/100mA
±(2%±20mA) 100mA ~ 40A
±(4%±0.3A) 40A ~ 65A
AC Current: 1mV/10mA
±(2%±30mA) 100mA ~ 10A (40Hz~ 2KHz)
±(4%±30mA) 100mA ~ 10A (2KHz~ 10KHz)
±(6%±30mA) 100mA ~ 10A (10KHz~ 20KHz)
±(8%±30mA) 10A ~ 15A (40Hz ~20KHz)
AC Current: 1mV/100mA
±(2%±30mA) 100mA ~ 40A (40Hz~ 1KHz)
±(4%±30mA) 100mA ~ 40A (1KHz~ 2KHz)
±(6%±30mA) 100mA ~ 40A (3KHz~ 5KHz)
±(8%±0.3A) 40A ~ 65A (40Hz ~20KHz)
Load Resistance: 10kΩ Typical
Temperature coefficient: 0.1×(definedaccuracy) per degree C (0℃to18℃,28℃to50℃)
SAFETY INFORMATION
The instrument complies with class II, overvoltage CAT II of the EN61010-1, and EN 61010-2-032 standards. The pollution of Current clamp is degree2 in accordance with IEC 664 indoor use. If the equipment is used in a mannernot specified, the protection provided by the equipment may be impaired.
This product complies with the requirements of the following EuropeanCommunity Directives: 89/336/EEC (Electromagnetic Compatibility) and 73/23/EEC(Low voltage) as amended by 93/68/EEC (CE marking).

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
Reply

Use magic Report

0

Threads

1

Posts

1

Credits

新手上路

Rank: 1

Credits
1
Post time 2017-7-25 01:33:17 | Show all posts
??????! ???????????????????????????????????? ?????????????????????????? ????????????????????????????????
www.justvirginhair.com/black-hairstyles.htm

0

Threads

2

Posts

2

Credits

新手上路

Rank: 1

Credits
2
Post time 2017-11-9 17:58:03 | Show all posts
We all wash our hair, but How Do You Wash your hair? Did you know there's a right way to wash your strands in the shower? Learning How To Wash Hair Properly will have a great impact on the bounce and beauty of your tresses. While most hairstylists recommend about a five-minute wash 3 to 4 times a week, I am willing to bet that most of you do something more like a one-minute wash. Am I right? Learn the proper steps to hair washing when you keep scrolling down.



You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-3-29 07:33 GMT+8 , Processed in 0.167629 second(s), 22 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list