Introduction to the Instrument

file:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml11136/wps1.pngfile:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml11136/wps2.pngfile:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml11136/wps3.pngfile:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml11136/wps4.pngFront Panel Review




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
Reply

Use magic Report

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
195#
 Author| Post time 2024-4-20 08:28:03 | Show the author posts only
Temperature Measurements

Temperature measurements require a temperature transducer probe. The supported probes are
2-wire and 4-wire RTDs, 2-wire and 4-wire thermistors.
Probe Type Choice
RTD's provide very accurate, highly linear relationships between resistance and temperature,
over a range of roughly –200 to 500 °C. There is very little conversion complexity for an RTD
because it is so intrinsically linear. The multimeter provides measurement for the IEC751 standard
RTD, which has a sensitivity of 0.385%/°C.


58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
194#
 Author| Post time 2024-4-19 08:06:10 | Show the author posts only
Avoid applying signals to unused current input terminals

If signal inputs are applied to terminals not needed for the current measurement, measurement
errors may occur. The unused terminals are still protected but the un-needed signals may interfere
with current measurement. For example, applying inputs to the 3A terminals while making
measurements on the 10A terminals will typically cause errors.

The Hi and Lo sense terminals are not used for many measurements. Applying signals here when
not needed can also cause errors. AC or DC voltages above 15 volts peak on the un-needed sense
terminals are likely to cause measurement errors. If unexpected errors are occurring, signals on the
un-needed terminals is an area to check.


58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
193#
 Author| Post time 2024-4-18 11:42:54 | Show the author posts only
DC Current

When you connect the multimeter in series with a test circuit to measure current, a measurement
error is introduced. The error is caused by the multimeter's series burden voltage. A voltage is
developed across the wiring resistance and current shunt resistance of the multimeter, as shown
below.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
192#
 Author| Post time 2024-4-16 08:07:20 | Show the author posts only
Other Primary Measurement Functions
Frequency and Period Measurement Errors
           
The multimeter uses a reciprocal counting technique to measure frequency and period. This method
generates constant measurement resolution for any input frequency. The multimeter's AC voltage
measurement section performs input signal conditioning. All frequency counters are susceptible to
errors when measuring low–voltage, low–frequency signals. The effects of both internal noise and
external noise pickup are critical when measuring "slow" signals. The error is inversely proportional
to frequency. Measurement errors also occur if you attempt to measure the frequency (or period)
of an input following a DC offset voltage change. You must allow the multimeter's input DC blocking
capacitor to fully settle before making frequency measurements.


58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
191#
 Author| Post time 2024-4-15 08:12:54 | Show the author posts only
This table gives an additional error for each waveform, to be added to the value from the accuracy
table provided in the instrument's data sheet.

The specifications are valid for CF ≤ 10, provided there is insignificant signal energy above the 300 kHz
bandwidth for voltage, or the 10 kHz bandwidth for current. Multimeter performance is not specified for
CF > 10, or when significant out-of-band signal content is present.
Example
A pulse train with level 1 Vrms, is measured on the 1 V range. It has pulse heights of 3 V (that is, a Crest Factor of 3) and duration 111 ?s. The prf can be calculated to be 1000 Hz, as follows:




file:///C:/Users/ADMINI~1/AppData/Local/Temp/ksohtml6328/wps3.png
Thus, from the table above, this AC waveform can be measured with 0.18 percent additional error.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
190#
 Author| Post time 2024-4-12 08:09:54 | Show the author posts only
The following table shows the typical error for various pulse waveforms as a function of input pulse
frequency.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
189#
 Author| Post time 2024-4-11 08:45:52 | Show the author posts only
Notice that crest factor is a composite parameter, dependent upon the pulse width and repetition
frequency; crest factor alone is not enough to characterize the frequency content of a signal.

Traditionally, DMMs include a crest factor derating table that applies at all frequencies. The
measurement algorithm used in the DMMs is not inherently sensitive to crest factor, so no such
derating is necessary. With this multimeter, as discussed in the previous section, the focal issue
is high–frequency signal content which exceeds the multimeter’s bandwidth.

For periodic signals, the combination of crest factor and repetition rate can suggest the amount of
high– frequency content and associated measurement error. The first zero crossing of a simple
pulse occurs at f1 = 1/tp .

This gives an immediate impression of the high-frequency content by identifying where this crossing
occurs as a function of crest factor: f1=(CF2)(prf).


58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
188#
 Author| Post time 2024-4-10 08:51:19 | Show the author posts only
Estimating High–Frequency (Out–of–Band) Error
A common way to describe signal waveshapes is to refer to their "Crest Factor". Crest factor is the
ratio of the peak value to rms value of a waveform. For a pulse train, for example, the crest factor
is approximately equal to the square root of the inverse of the duty cycle.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1267

Posts

1347

Credits

金牌会员

Rank: 6Rank: 6

Credits
1347
187#
 Author| Post time 2024-4-8 08:55:51 | Show the author posts only
In contrast, the frequency spectrum of the broad pulse has fallen off significantly below the multimeter’s
300 kHz (approximately) bandwidth, so measurements of this pulse are more accurate.

Reducing the prf increases the density of lines in the Fourier spectrum, and increases the portion of
the input signal’s spectral energy within the multimeter’s bandwidth, which improves accuracy.

In summary, error in rms measurements arise when there is significant input signal energy at
frequencies above the multimeter’s bandwidth.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-5-5 02:46 GMT+8 , Processed in 0.226804 second(s), 23 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list