Put the new fuse into the fuse holder and insert the assembly back into the instrument, turning the fuse holder clockwise to lock it in place.
If the 3 A path still has a bad fuse, replace the internal 3 A current path fuse as described below.
10 A Current Path To determine whether the instrument's 10 A current path has a fuse that must be replaced, press [Cont] to put the DMM in continuity measurement mode, and short the HI Input terminal to the 10 A current terminal. If the measurement displays OPEN, replace the internal 10 A current path fuse as described below.
|