One Channel of my Hantek IDS01070A S/N 612001429 has a defective OPAMP.
I cannot find the part number to replace the OPAMP.  Could you please tell me the manufacturer part number so I can replace the OPAMP?  Hope to hear from you soon.


Reply

Use magic Report

167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
2#
 Author| Post time 2020-12-14 15:10:59 | Show all posts
Reply

Please upload a picture to study.

167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
3#
 Author| Post time 2020-12-17 09:43:02 | Show all posts
Edited by amy at 2020-12-17 16:04

Sorry, I don't understand what you mean.
What's wrong with your oscilloscope? Cause you to suspect OPAMP's problem.
Where does the square signal in the picture come from?
Please use the oscilloscope to test the built-in square wave of the oscilloscope and upload the picture to study.

167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
4#
 Author| Post time 2020-12-17 16:06:14 | Show all posts
Edited by amy at 2020-12-17 16:30

Please set the probe attenuation coefficient of the software and the probes to ×1, and take a picture to study.

167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
5#
 Author| Post time 2020-12-18 14:07:50 | Show all posts
Do you have a signal generator?
Please input 1KH/1Vpp square wave to CH1 and CH2 at the same time, and set two channel to 100mV/div.
And upload a picture to study.

167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
6#
 Author| Post time 2020-12-19 08:21:47 | Show all posts
engineer@home replied at 2020-12-18 16:42
Dear Sir, Ma'am,

  Attached is the screen capture of the IDSO1070A connected to a signal generator ...

Sorry, please set the probe attenuation coefficient of the software and the probes to ×1, and take a picture again.

167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
7#
 Author| Post time 2021-2-22 16:35:07 | Show all posts
Sorry, there is no.
CC65 has two gears: 1:10 and 1:100, you can use ×10 or ×100 probe coefficient to test.
When calculating, we only need to replace V with A.