Hantek 6074BC faulty measurement.

[Copy link]

2

Threads

2

Posts

2

Credits

新手上路

Rank: 1

Credits
2
Jump to specified page
1#
Post time 2019-1-30 10:10:05 | Show the author posts only Reply Awards |Descending |Read mode
Edited by measureman at 2019-1-30 10:09

Hello

Square 32khz measurement is inaccurate.
Is it possible to have a faulty instrument ?

I've made the probe compensation, but haven't used it.

There is a similar error with a 555 timer ic.
Good at low sample speed.
There is an error at a higher sample speed.
Timer ic input voltage 8.00 v



What else can be a problem if the instrument is correct ?







This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
Reply

Use magic Report

167

Threads

2070

Posts

2232

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2232
2#
Post time 2019-2-14 15:39:12 | Show the author posts only
In the second picture, please set acquire->acquisition->peak detect, and take a picture to upload.
In the first picture, please set the volt/div to 2V and turn the capacitance on the probe to flatten square wave. Now test the fourth picture signal.