DSO4254B channel 2 Damaged?!?

[Copy link]

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
12#
Post time 2021-1-7 16:22:33 | Show the author posts only
Can I use 1Khz/5Vpp probe compensator from my other Hantek DSO5102P or must be 10Vpp?
With "their probe ratio is consistent" you mean the same (all 4ch x1 or x10)?

ThankYou. I will test as soon as I can

167

Threads

2069

Posts

2230

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2230
11#
 Author| Post time 2021-1-7 15:02:57 | Show the author posts only
Judging from the problem, it should be a hardware failure. Let's do some hardware tests. If you can solder, please follow these steps.

1. Open all channels, switch the volt/div between 200mV and 500mV, and check whether there is relay sound.
2. Input 1kHz/10Vpp square to four channels at the same time, their voltdiv is set to 100mV, their probe ratio is consistent. Compare and test the waveforms of several points marked in the figure below.


If you can't weld, please contact the seller for repair.

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
10#
Post time 2021-1-6 19:17:00 | Show the author posts only
amy replied at 2021-1-5 16:21
Reply

Do you swap probes to test ch1 and CH2? Does the oscilloscope still show CH2 abnormality?

can I send you some videos to caiyijun@hantek.com in order to study the problem?

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
9#
Post time 2021-1-5 21:30:45 | Show the author posts only
...and in the last screen with all 4 probes connected at 1Khz, 2Vpp only the waveform on channel 2 floats up / down on the screen. The other 3 are stable ...

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
8#
Post time 2021-1-5 19:12:21 | Show the author posts only
Only on ch2 ground is shifted...

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
7#
Post time 2021-1-5 18:59:22 | Show the author posts only
After further calibrations, the procedure was completed successfully. Now I get a ground shift on channel 2.
Both channel have x10 attenuation on probe.




This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
6#
Post time 2021-1-5 17:51:08 | Show the author posts only
More than 30 minutes have passed since it was turned on. I disconnected all the probes from the channels and performed a series of self-calibration.
I get always different errors.

Self Calibration 1
[all 4ch]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW then error 52102

Self Calibration 2
[all 4ch]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW then error 52126

Self Calibration 3
[all 4ch]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  then error 52122

Self Calibration 4
[all 4ch]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW
[1&2]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW
[3&4]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW then error 51201

Self Calibration 5
[all 4ch]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV then error 52131

Self Calibration 6
[all 4ch]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW
[1&2]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW
[3&4]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW
[1]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW
[2]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV
2mVBW  1mVBW 500uVBW then error 50100

Self Calibration 7
[all 4ch]
10V 5V  2V 1V 500mV 200mV 100mV 50mV 20mV 10mV 5mV then error 52132

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
5#
Post time 2021-1-5 17:03:06 | Show the author posts only
See also this...
ch2 with x1 and 5v arduino...

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
4#
Post time 2021-1-5 16:49:28 | Show the author posts only
see the trigger level on ch2.
I pushed 50% and DSO set to 6.2v. Why?
Label on probe comparator connector shows 1Khz 2Vp~p

Is there something inside ch2 that I can check? I've also another Hantek DSO (DSO5102P)

1

Threads

33

Posts

33

Credits

新手上路

Rank: 1

Credits
33
3#
Post time 2021-1-5 16:43:02 | Show the author posts only
same probe(x1) on both channel 1(yellow) & 2(blue)...

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-6-4 23:22 GMT+8 , Processed in 0.199325 second(s), 19 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list