Title: [SOLVED!!] DSO4072C Both channels BAD.
Edited by X-Blade at 2018-8-26 06:24

Hello.

I bought an hantek DSO4072C from a seller from ebay.
It was working good but after few days both channels went bad.
Seller dont answer my  messages and left me alone with the scope.

When I press Autoset button both channels turn on with a lot of noise, specially the CH2.

I can solder and replace components, just tell me what to replace and part numbers, if you can.
Can anybody help me, please?

Thank you.



Author: Newbee    Time: 2018-7-13 08:08
Hello  you can try this :Remove the probe    Press Utility button----Self Calibration---Confirm  then wait for a monent,  Restart  the machine. Good luck for you.

Sorry for my bad English

Author: X-Blade    Time: 2018-7-14 05:35
Thank you for your help Newbee.
Unfortunately I tried it but if hangs on 13/38 forever.
I will try to send it to somebody who represents Hantek.
This cheap chineese oscilloscopes are more a toy than a tool.
My hitachi V212 analog one resists to ervery abuse I subject it.

Author: amy    Time: 2018-7-19 09:07
X-Blade replied at 2018-7-14 05:35
Thank you for your help Newbee.
Unfortunately I tried it but if hangs on 13/38 forever.
I will try t ...

Sir, in your picture, did you input any signal to the channel?

Author: X-Blade    Time: 2018-7-20 05:16
Hello amy.
Yes is connected to the internal 1khz output on the panel.
It acts like there is no ground connected.
Very strange.
It started to do this after i was measuring an inductor SRF  without caution or some problem related to the common ground.

Author: X-Blade    Time: 2018-8-1 02:45
Edited by X-Blade at 2018-8-1 03:12

Hello friends.
Thank you for all your support.
I was searching another forum and I found the schematic (not official / Reverse-engineered) of the both Channel inputs:



In another thread I saw someone reporting a similar problem , so I replaced all the 5 sot-23 components of each channel:


BAV99 (marked A7), MMBTH10 (marked 3EM), MMBF4393 (Marked 6G) and the 2 x BC846B (marked 1B)

Now both channels are working but still not working correctly but exactly in the same way:
I connected the probe to the front calibration output (5v @ 1KHZ), probe at 1X and channel with 1x setting:


Looks good and accurate, but when I try to change the vertical position the amplitude of the waveform decreases and signal get distorted:


Interesting thing: I AM NOT ABLE TO BRING THE WAVEFORM BELOW THE  CENTER HORIZONTAL LINE FROM THE GRID BUT ABOVE IT MOVES CORRECTLY, expept the variation on the amplitude.
Both channels behave the same.
I have another oscilloscope (analog) I can perform some measurements.
Any suggestion?

Thank you all.





Author: X-Blade    Time: 2018-8-2 01:18
Hello people.

I found that -5V rail from 79M05 is putting out -7,4V (-5V Testpoint) . It looks like the regulator is gone, I will try to get a new one, replace and retest.
Cheers.

Author: X-Blade    Time: 2018-8-26 06:23
FINALLY it was solved.

Everyone can check my steps on
https://www.eevblog.com/forum/re ... -both-channels-bad/

Thank you.