The high impedance passive probe input impedance is high,
but its bandwidth is not high, low resistance passive probe
bandwidth can be high but the input impedance is not high.
Then is there a probe with both high input impedance and
high bandwidth? Actually yes, it is active probe.
In fact, active probe is a broader understanding, indicates
the probe need to be powered.
follows.
[attach]1638[/attach]
Author: WisdomAugust Time: 2017-7-17 08:29
There is a high-bandwidth amplifier on the front end of the active probe.
The amplifier needs power supply, and the name of active probe attribute
to it. The input impedance of the amplifier is relatively higher, so the
active probe can provide a relatively higher input impedance; while
the amplifier output drive capability is very strong, so it can directly
drive the rear 50 ohm load and transmission line. Because the 50-ohm
transmission line can provide a high transmission bandwidth, plus the
high bandwidth of the amplifier itself, so the entire probe system
compared to passive probes can provide higher bandwidth.
Author: WisdomAugust Time: 2017-7-18 16:16
We can see all the excellent features of the active probe are
brought by the front amplifier, but this high bandwidth amplifier
is expensive and need to be placed in a limited space in the
front of the probe, resulting in high cost.
General passive probe prices are hundreds of dollars or so,
while the price of active probes usually are thousands of dollars.
So the active probe is generally used in the need for the
occasion pf high measurement bandwidth.
Author: WisdomAugust Time: 2017-7-19 16:49
However, the active probe is not no shortcomings, limiting the wide
usage of active probes besides price factors, the other reason is its
limited dynamic range. The input voltage range like this high bandwidth
amplifier is not very large, so its measurement range can not be as big
as the passive probe. Commonly used 10: 1 passive probe, its dynamic
range usually at hundreds of volts, and it’s just a few volts for common
active probes. So there will be some limiation for its application.
Author: WisdomAugust Time: 2017-7-20 13:57
There is also a branch of the active probe, that is differential active probe.
The difference is that the front amplifier is a differential amplifier.
The advantage of the differential amplifier is that it can directly test the
high-speed differential signal, while its common mode rejection ratio is higher,
the common mode noise rejection ability is better.
[attach]1648[/attach]
Author: WisdomAugust Time: 2017-7-21 11:06
Current probe is another kind of active probe. There’s a magnetic ring
on the front end of the current probe, when measure, the magnetic ring
cover on the measured power line. Because the generated magnetic
field when the current flowing through the wire, which was collected by
the magnetic ring, the magnetic flux in proportion to the current flowing
through the wire.
Author: WisdomAugust Time: 2017-7-22 09:14
There’s a Hall sensor inside magnetic ring, which can detect
the magnetic flux and its output voltage in proportion to magnetic
flux. Therefore, the current probe output voltage is in proportion
to the measured current flowing through the wire. The conversion
factor for a typical current probe is 0.1V / A or 0.01V / A.
Author: WisdomAugust Time: 2017-8-3 08:47
The main advantage of the current probe is that it can measure current without
disconnecting the power line. Based on the Hall effect, it can measure both DC
and AC. Typical applications of current probes are system power measurement,
power factor measurement, and switching machine impact current waveform
measurements. The main drawback of the current probe is that its small current
measurement capability is limited by the bottom noise of the oscilloscope, so the
small current measurement capability is limited. Generally less than 10mA of
current is difficult to measure.
[attach]1651[/attach]