Title: How to select a probe?
Below is a list you need to consider when you choose a probe for a given measurement.

Input resistance

Input impedance is used to describe the loading effects of a probe. At DC and low frequency ranges, the probe's resistive component is the main factor that loads down the circuit under test. But as the frequency goes up, the capacitance of the probe tip, in parallel with the input impedance of the probe, resulting in greater loading and a more adverse effect to the target.

Author: WisdomAugust    Time: 2016-6-1 11:34
Input Capacitance
Excessive input capacitance shows down the system's pulse response. Usually the least input capacitance possible is best.

Author: WisdomAugust    Time: 2016-6-2 12:25
Maximum input voltage
To ensure user safety, help protect the oscilloscope input from destructive voltage, and avoid damage to the probe, select a probe that is rated for a higher voltage than the signal you intend to test.

Author: WisdomAugust    Time: 2017-4-20 09:29
Bandwidth
The rated bandwidth of a probe should match the one of oscilloscope and be adequate for the test signal.But at higher frequencies, grounded lead inductance and input capacitance often influence system performance more than probe bandwidth.

Author: WisdomAugust    Time: 2017-5-3 15:56
Attenuation
Choose the attenuation ratio of the probe (X1, X10, X100, X1000) to match the test signal amplitude to the vertical sensitivity range of the oscilloscope.

Author: WisdomAugust    Time: 2017-5-4 16:29
Probe compensation range

Most passive probes have a specification that lists the oscilloscope input capacitance
range over which they can be used. When choosing a passive probe, be sure that the
oscilloscope's input capacitance lies within the probe's compensation range or you will
not be able to adjust the probe to chieve a correctly compensated square wave signal.



Author: WisdomAugust    Time: 2017-5-5 15:25
Edited by WisdomAugust at 2017-5-5 15:27

Most oscilloscopes have 1 ohm input resistance. This input resistance
is in parallel with the input capacitance. Normally, high-frequency probes
with attenuation factors grater than X1 have adjustable compensation
networks built into them. Adjusting this compensation network provides
the best possible frequency linearity over the oscilloscope's desinged
frequency range. Operating instructions provided with probe explain
how to adjust the compensation network to obtain best signal fidelity.

[attach]1506[/attach]

Author: WisdomAugust    Time: 2017-5-8 11:47
To select the appropriate probe, the first thing to do is to understand
the impact of the probe on the test, which includes the meaning of 2 parts:
1) the impact of the probe on the to be tested circuit;
2) The signal distortion caused by the probe.
The ideal probe should have no effect on the to be tested circuit, meanwhile
no distortion on the signal. Unfortunately, there is no such a probe to meet
these two conditions at the same time, usually need to do some trade-off
between the two parameters.


Author: WisdomAugust    Time: 2017-5-9 10:41
In order to consider the impact of the measurement of probe,
usually we simply take the probe model as a R, L, C model,
and analysis the model and to be tested circuit together.

[attach]1516[/attach]


Author: WisdomAugust    Time: 2017-5-10 13:22
First, the probe itself has an input resistance. The same theory as
testing voltage by multimeter, in order to minimize the impact of
the to be tested circuit, requires the input resistance (Rprobe) of
the probe itself as much as possible. However, because it's impossible
for Rprobe to be infinity, shunting will be generated with the to be
tested circuit, and the actual measured voltage may not be the
the real voltage on probe point, which is ofter encountered on some
power or amplifier circuit test. To avoid the impact of the load resistance
of the probe, there is the level required of Rprobe-10 times more than
Rsource and Rload. Most of the probe's input impedance is between
10K ohms and 10M ohms.


Author: WisdomAugust    Time: 2017-5-15 09:00
Second, the probe itself has an input capacitor, which is not deliberately done, but is parasitic.
This parasitic capacitance is also the most important factor affecting the bandwidth of the probe,
because this capacitor will attenuate high frequency, and slow the rising edge of the signal.
Usually high-bandwidth probe parasitic capacitance is relatively small. Ideally the Cprobe should
be 0, but actually it can not. The input capacitance of the common passive probe is between
10pf to several hundreds of pf, and the input capacitor of higher bandwidth active probe
is usually between 0.2pf and several pf.


Author: WisdomAugust    Time: 2017-5-16 16:32
Third, the probe input will be affected by the inductance. The input resistance
and capacitance of the probe are better to be understood, but  the inductor at
the input of the probe is often overlooked, especially when do high frequency
measurements. Where the inductance comes from? We know where is a wire,
there is a inductor. Between the probe and the measured circuit there must be
a connecting wire, meanwhile the return of the signal also passes through the
ground of the probe.



Author: WisdomAugust    Time: 2017-5-17 08:55
Usually 1mm probe ground will own about 1nH inductance. The longer signal and ground,
the greater inductance. The parasitic inductance and parasitic capacitance of the probe
composed a resonant loop. If the inductance value is too big, it is possible to generate
high-frequency resonance under the excitation of the input signal, resulting in distortion
of the signal. So the high-frequency testing requires strict control of the length of the
signal and ground, otherwise it is prone to ringing.
[attach]1518[/attach]


Author: WisdomAugust    Time: 2017-6-2 08:12
Edited by WisdomAugust at 2017-6-2 08:13

Before learning the structure of the probe, we need to know the structure
of the oscilloscope input interface first, because it is the place where the
probe is connected. The input interface circuit oscilloscope and the probe
together composed the detection system.

[attach]1528[/attach]




Author: WisdomAugust    Time: 2017-6-3 08:56
Most of the oscilloscope input interfaces are in the form
of BNC or BNC compatible. The input of the oscilloscope
has a matching resistance of 1M ohms or 50 ohms. There
are many types of oscilloscope probes, but the oscilloscope's
match is always 2 options: 1M ohm or 50 ohm. Different types
of probes need different matching resistance.


Author: WisdomAugust    Time: 2017-6-5 08:41
From the voltage measurement perspective, to minimize the
impact of to be measured circuit, the oscilloscope can use
a high input impedance of 1M ohm. However, the bandwidth
of the high impedance circuit is susceptible to parasitic
capacitance, so 1M ohm input impedance is widely used with
measurements below 500M bandwidth. For higher frequency
measurements, a 50 ohm transmission lead is usually used,
so the 50 ohm match of the oscilloscope is mainly used for
high frequency measurements.



Author: WisdomAugust    Time: 2017-6-6 11:45
Edited by WisdomAugust at 2017-6-6 11:54

Traditionally, most of the oscilloscopes below the 100MHz bandwidth
on the market have only 1M ohm inputs because they are not used for
high frequency measurements; most of the 100MHz to 1GHz bandwidth
oscilloscopes have both 1M and 50 ohm switching options, taking into
account both high and low frequency measurements; 2GHz or higher
bandwidth oscilloscopes are mainly used for high-frequency measurement,
so most of them only have 50 ohm input. However, in view of the needs of
the market, some oscilloscopes above 2GHz also provide 1M ohm and 50
ohm input switch.



Author: WisdomAugust    Time: 2017-6-7 08:16
Broadly speaking, the test leads belongs to probe as well,
like BNC or SMA cable. This type of probe is cheap and high
performance. But the use of test leads connection require the
to be tested circuit has BNC or SMA interface, thus their application
is limited, mainly for radio frequency and microwave signal testing.
For digital or general signal testing, should use a special probe.



Author: WisdomAugust    Time: 2017-6-10 10:33
Edited by WisdomAugust at 2017-6-22 13:20

According to the power supply oscilloscope probe can be
divided into passive probe and active probe, sorted by
the measured signal type the probe can be divided into
voltage probe, current probe, optical probe etc. The
so-called passive probe indicates the entire probe by
composed by passive components, including resistors,
capacitors, cables etc; well the active probe inside
has general amplifier, and the amplifier need to be
powered, so named active probe.


Author: WisdomAugust    Time: 2017-6-12 08:27
Edited by WisdomAugust at 2017-6-22 13:21

Sorted by  the size of the input impedance is divided into
two types: high impedance passive probe and low resistance
passive probe.

High-impedance passive probe is the one that we usually call
passive probe, which is the most widely used probe, basically
every person using the oscilloscope have used this type of probe.
When the high-impedance passive probe is connected to the
oscilloscope, the input impedance of the oscilloscope is
required to be 1M ohms. The following is a schematic diagram
of a 10: 1 high impedance passive probe.

[attach]1595[/attach]

Author: WisdomAugust    Time: 2017-6-15 08:27
To facilitate the measurement, the probe usually
has around 1 meter length. If not adding the matching
circuit, it is hard to imagine that the probe can
provide hundreds of megabytes of bandwidth. The input
parasitic capacitance of the oscilloscope affects the
bandwidth as well.

To improve the high frequency corresponding of the probe,
there will be a corresponding matching circuit on the
probe’s front-end. The most typical is a Rprobe and
Cprobe parallel structure.


Author: WisdomAugust    Time: 2017-6-16 08:10
A condition for the probe to produce a flat gain in
the band is to meet Rprobe * Cprobe = Rscope * Cscope.
Cscope is the parasitic capacitance of the oscilloscope,
so it can only be controlled in a certain range,but
can not be precisely controlled, i.e. The Cscope value
of different oscilloscopes or different channels on
oscilloscope will be different. In order to compensate
for changes in different channels Cscope, at the end of
the probe connected to the oscilloscope, where will be
at least an adjustable capacitor-Ccomp.


Author: WisdomAugust    Time: 2017-6-19 08:24
Edited by WisdomAugust at 2017-6-22 13:21

When the probe is connected to different channel of
the oscilloscope, it can compensate Cscope changes
by adjusting the Ccomp. Almost all oscilloscopes
provide a low frequency square wave output,
which can connect to the probe to adjust
the shape of the saqure wave to adjust
the Cscope.

[attach]1596[/attach]

Author: WisdomAugust    Time: 2017-6-20 09:21
Rprobe improves the frequency response and produces a
divider with oscilloscope input resistance. The so-called
10: 1 divider is the actual measured voltage of the
oscilloscope is 1/10 of the voltage at the front end
of the probe, i.e. the signal through the probe will
have a 10 times attenuation.


Author: WisdomAugust    Time: 2017-6-21 08:11
Edited by WisdomAugust at 2017-6-22 13:22

The simpler probe needs to set the oscilloscope's attenuation manually
to get the correct display. More probes have an auto-detect pin at the
end which connect to the oscilloscope. It can read the attenuation
ratio of the probe through the pin when the probe is plugged in
the oscillsocope, and adjust the display ratio automatically.

[attach]1597[/attach]

Author: WisdomAugust    Time: 2017-6-22 13:17
There are two special types of high impedance passive probes.
One is a high-voltage probe, the attenuation ratio up to 100:
1 or 1000: 1, so the measurement voltage range is large;
another is a type of 1: 1 probe, i.e., there’s no attenuation
of the signal when input the oscilloscope. Not like 10: 1
probe which need the oscilloscope to enlarge the display,
so the noise of the oscilloscope itself is not amplified,
and the measured of noise will be much smaller, which use
a lot in the small signal and power ripple measurement.


Author: WisdomAugust    Time: 2017-7-12 14:43
Edited by WisdomAugust at 2017-7-12 14:49

Another passive probe is a low-resistance passive probe,
which is the least commonly used probe.
The biggest advantage of the low-resistance passive probe
is that it has relatively high test bandwidth (1G-6GHz) with
the price of an ordinary high-impedance passive probe.
The main drawback is that the input impedance is low
(only 500 ohm or 1k ohm), if connect to the circuit, it may
still have a little impact on the measured signal.
The following is its duagrammatic layout.

[attach]1630[/attach]



Author: WisdomAugust    Time: 2017-7-13 09:11
The low-impedance passive probe requires an oscilloscope's input impedance
of 50 ohms and its front-end equivalently series connect a divider resistor.
According to the different series resistance to achieve different partial voltage
ratio. For example, if series a 450 ohm resistor, the partial voltage is 10: 1.
As a result of 50 ohm transmission cable, and the oscilloscope is 50 ohm
as well, so the entire probe bandwidth is relatively high.
[attach]1636[/attach]


Author: WisdomAugust    Time: 2017-7-14 08:24
The high impedance passive probe input impedance is high,
but its bandwidth is not high, low resistance passive probe
bandwidth can be high but the input impedance is not high.
Then is there a probe with both high input impedance and
high bandwidth? Actually yes, it is active probe.
In fact, active probe is a broader understanding, indicates
the probe need to be powered.
The schematic diagram of porpular active voltage probes as
follows.

[attach]1638[/attach]



Author: WisdomAugust    Time: 2017-7-17 08:29
There is  a high-bandwidth amplifier on the front end of the active probe.
The amplifier needs power supply, and the name of active probe attribute
to it. The input impedance of the amplifier is relatively higher, so the
active probe can provide a relatively higher input impedance; while
the amplifier output drive capability is very strong, so it can directly
drive the rear 50 ohm load and transmission line. Because the 50-ohm
transmission line can provide a high transmission bandwidth, plus the
high bandwidth of the amplifier itself, so the entire probe system
compared to passive probes can provide higher bandwidth.


Author: WisdomAugust    Time: 2017-7-18 16:16
We can see all the excellent features of the active probe are
brought by the front amplifier, but this high bandwidth amplifier
is expensive and need to be placed in a limited space in the
front of the probe, resulting in high cost.
General passive probe prices are hundreds of dollars or so,
while the price of active probes usually are thousands of dollars.
So the active probe is generally used in the need for the
occasion pf high measurement bandwidth.


Author: WisdomAugust    Time: 2017-7-19 16:49
However, the active probe is not no shortcomings, limiting the wide
usage of active probes besides price factors, the other reason is its
limited dynamic range. The input voltage range like this high bandwidth
amplifier is not very large, so its measurement range can not be as big
as the passive probe. Commonly used 10: 1 passive probe, its dynamic
range usually at hundreds of volts, and it’s just a few volts for common
active probes. So there will be some limiation for its application.


Author: WisdomAugust    Time: 2017-7-20 13:57
There is also a branch of the active probe, that is differential active probe.
The difference is that the front amplifier is a differential amplifier.
The advantage of the differential amplifier is that it can directly test the
high-speed differential signal, while its common mode rejection ratio is higher,
the common mode noise rejection ability is better.
[attach]1648[/attach]


Author: WisdomAugust    Time: 2017-7-21 11:06
Current probe is another kind of active probe. There’s a magnetic ring
on the front end of the current probe, when measure, the magnetic ring
cover on the measured power line. Because the generated magnetic
field when the current flowing through the wire, which was collected by
the magnetic ring, the magnetic flux in proportion to the current flowing
through the wire.


Author: WisdomAugust    Time: 2017-7-22 09:14
There’s a Hall sensor inside magnetic ring, which can detect
the magnetic flux and its output voltage in proportion to magnetic
flux. Therefore, the current probe output voltage is in proportion
to the measured current flowing through the wire. The conversion
factor for a typical current probe is 0.1V / A or 0.01V / A.


Author: WisdomAugust    Time: 2017-8-3 08:47
The main advantage of the current probe is that it can measure current without
disconnecting the power line. Based on the Hall effect, it can  measure both DC
and AC. Typical applications of current probes are system power measurement,
power factor measurement, and switching machine impact current waveform
measurements. The main drawback of the current probe is that its small current
measurement capability is limited by the bottom noise of the oscilloscope, so the
small current measurement capability is limited. Generally less than 10mA of
current is difficult to measure.
[attach]1651[/attach]