Electronic Engineer Discuss

Title: Hantek HT824 - Process Calibrator [Print this page]

Author: CYH2018    Time: 2022-12-30 08:58
Title: Hantek HT824 - Process Calibrator
Customer service,

We have an HT824 Process Calibrator that we have checked and is out of calibration.  I’ve attached the datasheet for reference.  Can you help me with the process of calibrating or exchanging it since it is still under warranty?

HT824 - Process Calibrator
Serial Number HT160100316
Purchase Date: 2022-01-19

Horner APG
59 S. State Ave
Indianapolis, In 46201


Thanks,
Brian


Author: CYH2018    Time: 2022-12-30 09:53

Dear Sir,

Greeting from Hantek.

We can give you free calibration, but you need to pay for the round-trip postage and customs fees.
The procedure for calibration is provided in the manual, or you can perform the calibration yourself.
I will send you the calibration manual.

Have a nice day.


Author: CYH2018    Time: 2023-1-29 09:14
Thank you for the calibration manual.  It is very helpful.  Stepping through the process works very well until I get to the following:

○ Press ENTER and CAL_DOWN keys at the same time to make the monitor display "SCAL" - millivolt output full-scale calibration.
○ If the value displayed on the monitor does not match the reading of the standard meter (-10.000mV), press the direction down key (the reading of the standard table will become smaller) and press the direction up arrow key (the reading of the standard table will become larger). Adjust until the two are equal.

The source reading on the Hantek is -10.000mV SCAL and my monitor -10.136.  When I press the up arrow the reading doesn’t change, but pressing the down arrow causes it to go in the wrong direction.   It changes from      -10.136, -10.157, -10,162….  Do you have any suggestions?

Author: CYH2018    Time: 2023-1-29 09:43
Dear Sir,

Greeting from Hantek.

Is it possible that the up arrow is broken, or you are advised to operate several times.

Have a nice day.

Author: HornerAPG    Time: 2023-4-28 20:47
CYH2018 replied at 2022-12-30 09:53
Dear Sir,

Greeting from Hantek.

Can you tell me the procedure to follow to return the device for calibration under warranty?

Thanks





Welcome to Electronic Engineer Discuss (https://www.eediscuss.com/) Powered by Discuz! X3.2