My brand new 2D10 shows a strange "interference signal" when the memory depth is set to more than 40k. The interference signal always appears with a time delay after increasing the memory depth and overlays the measurement signal. (Many, very short spikes on the screen) It also appears without any probes connected. Can I regard this behaviour as another bug or is there possibly a hardware defect?
I have already tested different software and Firmware versions. (At this moment, the actual Firmware from December 30. 2021 is installed, it doesn't help...)
Here a video of this issue.
https://1drv.ms/v/s!AiPaG0Dlc3q2gZZ0pJNZPpmzM4aY8g?e=anox9T









Author: naelolaiz    Time: 2022-1-5 04:48
Edited by naelolaiz at 2022-1-5 06:37

I have the same issue with a brand new 2D15. I also tested it with the provided firmware, and the latest one.
Here are my videos:
https://www.youtube.com/watch?v=7vd8mAi3oPI
https://www.youtube.com/watch?v=MJYGLvc7Mws

Edit: other issues on the device:First: the sine frequency does not match the frequency of the generator.  Or the generator is really that bad that makes a 0.6 Hz modulation? Second: a different type of noise appears at the end of the buffers acquiring at 40K.I don?t trust this device as a measurement/testing equipment as far as it is behaving. Is is expected to behave this bad? Are all these hardware or software/firmware issues?
video url: https://www.youtube.com/watch?v=186fGh3LLVw
                                                        
                                            





Author: Maxie    Time: 2022-1-7 01:15
Unfortunately I have the same issue on my 2D10 I received just a couple of days ago from Banggood.
Displaying signal at 4k and 40k points works OK, but when I go higher in memory settings, the scope starts behaving badly.
Memory settings 400k, 4M and 8M are unusable.
This happens almost instantly when I change the Volts/div setting, zooming in on the signal.
If I switch it OFF and ON again, the problem comes back by adjusting V/div.
The problem exist also in GND coupled mode or even the probe disconnected with no signal at all.
Serial: CN21350290127xx
SW: 20210726
HW: 3.2.1.0.0.0.0.1
FW: 3202
I have since tried the latest SW from the Hantek site 20211230, it behaves the same.
I hope Hantek will reply, I'd like to know if it is a hardware or software defect and can we expect a fix?
My video of the problem is available here:
https://www.youtube.com/watch?v=gFO4QoYVsSE


Author: wangshuang    Time: 2022-1-7 11:57
naelolaiz replied at 2022-1-5 04:48
I have the same issue with a brand new 2D15. I also tested it with the provided firmware, and the la ...

Thank you for your feedback. This problem has been reported to the engineer, and I will notify you when a new firmware is released.

Author: masta-k74    Time: 2022-1-9 18:10
wangshuang replied at 2022-1-7 11:57
Thank you for your feedback. This problem has been reported to the engineer, and I will notify you ...

I assume now that you are also assuming a firmware bug and not a hardware defect?

Author: filipj    Time: 2022-1-13 01:31
I have also faced a problems with Memory depth larger than 4k, but with FFT (spectre on negative frequencies...)
https://www.eediscuss.com/forum. ... id=15885&extra=
Have not got any reply on this topic though...

Author: masta-k74    Time: 2022-1-21 13:52
Edited by masta-k74 at 2022-1-21 13:54

There is a new Firmware (dso3kb_20220118.upk) on official Website


Author: masta-k74    Time: 2022-1-26 03:27
This problem still persists...

Author: masta-k74    Time: 2022-2-3 20:32
One could almost get the impression that nobody is interested. I am well on the way to turning my back on Hantek for good. At any rate, you can't call that customer service. If a solution to the problem is still being sought, it would be desirable that it is at least communicated.
Return shipping to China is out of the question for cost reasons. Shipping from Germany costs about ?100, half the price of the device. In addition, you can read here from others who have tried it that there are sometimes problems with customs. Then the device and the money would be gone.


Author: Maxie    Time: 2022-2-9 00:55
Looks like there is no solution to this issue.
The only feedback we received is that the problem has been reported to the engineer.
Shall I return my scope and ask for refund?

Author: Vagrex    Time: 2022-2-14 20:26
The latest update didn't fix the problem. i'm afraid that it's hardware problem. Is there anybody there to inform us about the problem? we are so many people here waiting for answer. Thanks.

Author: PY2ZFC-CT1EFF    Time: 2022-2-18 02:02
Vagrex replied at 2022-2-14 20:26
The latest update didn't fix the problem. i'm afraid that it's hardware problem. Is there anybody th ...

My DSO2D15 purchased in October 2021, the memory depth works perfectly well in all situations 4K 40K 400K 4M 8M this with all software, came from factory with dso3kb_20210814.upk software, currently is with dso3kb_20220210.upk and the firmware remains the same 3202, I believe your DSO may have ardware problem.
From: Fernando Santos

Author: Jing    Time: 2022-2-18 16:52
Maxie replied at 2022-1-7 01:15
Unfortunately I have the same issue on my 2D10 I received just a couple of days ago from Banggood.
D ...

Please provide your email, or send me a message, support@hantek.com.

Author: Vagrex    Time: 2022-2-18 18:56
PY2ZFC-CT1EFF replied at 2022-2-18 02:02
My DSO2D15 purchased in October 2021, the memory depth works perfectly well in all situations 4K 4 ...

Τhe problem does not always occur. Have you tried to play with the volts/div knob while you are at 8M? It may take several minutes for the problem to occur but when it does it appears it is visible at 400k 4M and 8M. If you turn off and start the dso it disappears and it will take time for it to reappear. If you do not use any buttons and just have a waveform on the screen, it never does the problem. I don't think the volts/div knob is the problem. Whenever you make any change when you are in 4M or 8M(change input let's say) the problem may be visible.

Author: PY2ZFC-CT1EFF    Time: 2022-2-19 06:30
Vagrex replied at 2022-2-18 18:56
Τhe problem does not always occur. Have you tried to play with the volts/div knob while you are a ...

Vagrex Post!
Τhe problem does not always occur. Have you tried to play with the volts/div knob while you are at 8M? It may take several minutes for the problem to occur but when it does it appears it is visible at 400k 4M and 8M. If you turn off and start the dso it disappears and it will take time for it to reappear. If you do not use any buttons and just have a waveform on the screen, it never does the problem. I don't think the volts/div knob is the problem. Whenever you make any change when you are in 4M or 8M(change input let's say) the problem may be visible.

My answer to Vagrex!
Hi Vagrex!
After reading your post, I went to do the test as you reported, the test was exhausting, it took an hour, making changes to the memory depth, in VOLTS/DIV, SEC/DIV and TRIGGER LEVEL, it always worked correctly, without any problem. The serial number of my DSO2D15 is: CN2136029013321
From: Fernando Santos

Author: Vagrex    Time: 2022-2-19 13:30
PY2ZFC-CT1EFF replied at 2022-2-19 06:30
Vagrex Post!
Τhe problem does not always occur. Have you tried to play with the volts/div knob wh ...

Hi PY2ZFC-CT1EFF. Thanks for your try.
This is good news for you. For those of us who have the problem, can someone from Handek tell us what to do? To return the unit; No one answers.

Author: Maxie    Time: 2022-2-21 09:25
Edited by Maxie at 2022-4-27 15:57
Jing replied at 2022-2-18 16:52
Please provide your email, or send me a message, .

Hello Jing,

I have sent you an email as per your request.


Author: Maxie    Time: 2022-4-25 08:12
Just a quick feedback from me, the received FW upgrade fpga3203 solved the issue for me, I have not experienced the signal spikes any more.
I have also shared the package on the corresponding EEVBLOG forum page.


Author: mpaultech@gmail.com    Time: 2022-5-9 23:46
I purchased hantek DSO 2d10 has a problem of auto set button not functioning properly it doesn't give correct measurements when am TESTING AC SIGNALS. BECAUSE WHEN I PRESS AUTO SET BUTTON TO CAPTURE AC DC SIGNAL OR ANY Signal it doesn't CAPTURE AND SOMETIMES HANGS AND FREEZ. PLEASE I NEED SOLUTION FOR THIS CASE

Author: mpaultech@gmail.com    Time: 2022-5-9 23:47
AUTO SET BUTTON DEOSNT FUNCTION WHEN CAPTURING SIGNALS  (HANTEK 2D10