Author: wangshuang    Time: 2021-12-28 11:31
Title: 8060 stopped starting
Customer consultation

The oscillograph stopped starting - "Self check" does not pass. That is, after switching on - a short buzzer - 5 seconds of waiting - a click of the relay - switching off. It does not reach the start of the display.



Author: wangshuang    Time: 2021-12-28 11:40



Reply
You can measure the voltage at these points in the figure below.





Author: iMaks-RS    Time: 2021-12-28 15:02
my board is different

Author: wangshuang    Time: 2021-12-30 14:22
iMaks-RS replied at 2021-12-28 15:02
my board is different

Please measure these points in the figure below.


Author: iMaks-RS    Time: 2022-4-4 19:35
The voltages at the indicated points and the frequency are the same as those indicated in the figure (photo) 2-1.png. But - are present only after pressing the power button of the oscilloscope. They disappear after 3-5 seconds.
Figure (photo) 2-2.png is different from what I have.. the tantalum capacitor is missing. Tell me on what conclusions to take measurements. I put the voltmeter probe on the test point, the track from which leads to pin 2 of the U26 chip. There 0V.