Title: DSO4202C DC 3 mV offset when it warms up [Print this page] Author: wjw Time: 2019-12-5 04:18 Title: DSO4202C DC 3 mV offset when it warms up I have a DSO4202C.
After the unit warms up, there is a 3 to 5 mV DC shift in the measured voltage when no probe is connected or if the input is shorted external to the unit.
When I select GND from the CH1 or CH2 settings, it *does* go to zero.
It would be helpful if there was a way we could adjust the input settings to compensate for this error. I know I can change the level and shift it down, but then the measurements (the numbers displayed on the screen) are incorrect.
Author: wangshuang Time: 2019-12-5 11:50
Please perform Self Calibration. Author: Oleg10011001 Time: 2019-12-7 15:32
I should add that between turning on the oscilloscope and starting the "Self calibration" procedure, it should take 20 minutes for the oscilloscope to warm up. Author: wangshuang Time: 2019-12-10 16:47
What is the firmware version of the oscilloscope? Can you upload an operation video so we can reproduce the problem?
Welcome to Electronic Engineer Discuss (https://www.eediscuss.com/)