Title: "self calibration failed, error x10b00" [Print this page] Author: alper Time: 2019-7-26 04:19 Title: "self calibration failed, error x10b00" hi I received mso5202d calibration while "Self calibration failed, error 10b00" error and I can not use the second channel.Firmware file can not find.Please help. Author: amy Time: 2019-7-26 14:57
Please describe the specific problems with Channel 2.
It would be better to upload pictures. Author: alper Time: 2019-7-26 17:49 Edited by alper at 2019-7-27 21:51
calibrated error and the line of the second channel has disappeared
Data error:
Author: alper Time: 2019-7-30 04:00
can someone help me? Author: amy Time: 2019-7-30 11:16
The hardware should be fault.
Would you like to deal with it yourself first?
If not, please contact your seller to return the mainboard to factory for repair. Author: alper Time: 2019-7-30 14:45
I unpacked and calibrated without measuring.maybe if I install the software it will improve but I can't find the software. Author: alper Time: 2019-7-31 16:22
Does service@hantek.com.cn work? because they don't answer my mail. Author: Han Time: 2019-7-31 16:26
Yes,they will reply Author: amy Time: 2019-8-1 08:07 Edited by amy at 2019-8-1 08:13
This is not a firmware issue.
Upgrading can't solve the problem.
Because you have been answered in this forum, so the mail did not reply to you.
Author: alper Time: 2019-8-1 21:42
what can i do to solve the problem?
Author: amy Time: 2019-8-2 08:48
Please test these six points voltage.1 2.5V
2 -2.88V
3 1.8V (1 and 8 pins of U203)
4 +5V/-5V (6 and 7 pins of U203)
[attach]3043[/attach]
Author: alper Time: 2019-8-4 02:37 Edited by alper at 2019-8-4 02:42
Please test these six points voltage.1 2.49V
2 -2.97V
3 1.pin=2.40 and 8.pin=2.41 of U203)
Author: amy Time: 2019-8-5 11:48
Please test the voltage of this magnetic bead. It should be 12V.
[attach]3055[/attach]
Author: alper Time: 2019-8-6 04:30
"Please test the voltage of this magnetic bead. It should be 12V."
11.98V
Author: alper Time: 2019-8-6 05:45 Edited by alper at 2019-8-6 05:48
I think I found the fault.I removed D02_1 diode and voltage values returned to normal,ch2 line came back. But I did not have a full test because I do not have bav99 in my hand. I ordered the diode I will share the result with you Author: amy Time: 2019-8-6 14:36
Ok. Hope for good results. Author: alper Time: 2019-8-14 03:47
hi amy i changed d02_1 diod and everthing is ok. Ch_2 is working and self calibration ok.Thank you for help.