Electronic Engineer Discuss

View: 8955|Reply: 5
Print Prev. thread Next thread

Voltage calibration

[Copy link]

1

Threads

3

Posts

3

Credits

新手上路

Rank: 1

Credits
3
Jump to specified page
1#
Post time 2016-11-5 14:05:07 | Show the author posts only Reply Awards |Descending |Read mode
Hello I have brand new DSO5072P.
I run the self calibration a few times after unit has been on for some time.
When i connect the probes to the 5V calibration port they read 1khz perfectly but 5.2volt for the top voltage Vmid is 2.4 Volts and peak to peak voltage is 5.4 volts. . I have also tried many different probes and all show same voltage,
My fluke DMM shows the signal to be 2.5volts as expected and the calibration point.

How can I calibrate this scope to read voltage with accuracy, Can it be done in software?
Thank you.

Reply

Use magic Report

167

Threads

2069

Posts

2230

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2230
2#
Post time 2016-11-7 09:19:53 | Show the author posts only
Please upload a screenshot to study.

1

Threads

3

Posts

3

Credits

新手上路

Rank: 1

Credits
3
3#
 Author| Post time 2016-11-8 07:59:17 | Show the author posts only
Probe connected to calibration point.

Values is 5.20  Mean 2.48.

My fluke DMM shows perfect 2.5 the Calibration port is correct, the scope shows wrong voltage, is there method to calibrate it to 2.5 mean and 5 pk-pk

This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

167

Threads

2069

Posts

2230

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2230
4#
Post time 2016-11-8 11:55:46 | Show the author posts only
The oscilloscope DC Gain Accuracy is ±(3%*range) for Normal or Average acquisition mode, with 5V/div to 10mV/div.
In your picture, the Accuracy is ±0.48V.
You can change the voltage scale to 1V/div to measure.
Or click "Acquire->Average mode" to open average acquire mode to reduce the noise.

1

Threads

3

Posts

3

Credits

新手上路

Rank: 1

Credits
3
5#
 Author| Post time 2016-11-9 04:35:00 | Show the author posts only
Is there a way to offset the voltage, manually in the DSO firmware? or hidden menu?

167

Threads

2069

Posts

2230

Credits

版主

Rank: 7Rank: 7Rank: 7

Credits
2230
6#
Post time 2016-11-9 09:57:18 | Show the author posts only
theabyss replied at 2016-11-9 04:35
Is there a way to offset the voltage, manually in the DSO firmware? or hidden menu?

Sir, there is no method to offset the voltage.
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-6-15 20:25 GMT+8 , Processed in 0.221954 second(s), 22 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list