The DC voltage for the DMM portion is reading 5.01v when it should be reading 5.00v. I have a DMM calibrated DC source, and also comparing it to a known good Fluke DMM. I realize this is not a large difference, but it would be beneficial to have the 2D72 DMM closer calibrated. I have only tested the 2D72 DMM DC voltage functions so far. The Scope portion appears to work well. Is ther an adjustment port on the 2D72's PCB that I could adjust to bring the meter into better calibration?
Thank-you.
Lane replied at 2021-3-11 01:28
The DC voltage for the DMM portion is reading 5.01v when it should be reading 5.00v. I have a DMM ca ...
The DC voltage for the DMM portion is reading 5.01v when it should be reading 5.00v. I have a DMM calibrated DC source, and also comparing it to a known good Fluke DMM. I realize this is not a large difference, but it would be beneficial to have the 2D72 DMM closer calibrated. I have only tested the 2D72 DMM DC voltage functions so far. The Scope portion appears to work well. Is ther an adjustment port on the 2D72's PCB that I could adjust to bring the meter into better calibration?
Thank-you.
This post contains more resources
You have to Login for download or view attachment(s). No Account? Register
Lane replied at 2021-3-11 13:57
Yes, I had realized the specifications, but is there an adjustment pot on the PCB that could be used ...
Sorry, it cannot be done.
If 5V is calibrated, the test deviation of other voltages will be larger.
In addition, this DMM has a resolution of 0.01V in the 40V range, not 0.001V. It depends on the hardware.