So I am able to identify the problem and also solve it, although not very much sure how I did that, but I will share here.
When I buy this model Hantek 6022BL, I know there is a logic analyzer and oscilloscope inbuilt, but was not aware of the button high and low concept, i.e. High means use it as Logic Analyzer and Low means use it as Oscilloscope.
With Hantek provided software, both works fine at a time, of course after the installation of drivers.
Now let me share what exactly I did wrong.
When I received the Hantek 6022BL, the button was in P (un-pressed) position, and showing Salae Logic Analyzer, at that time I was not aware of the fact that we can use this device with Salae software, so I thought why the hell this is showing this and I uninstalled the driver.
And then I installed the Hantek Drivers and it was working fine for both mode i.e. Oscilloscope and Logic Analyzer, but this causes Blue Screen multiples times on my computer and in that Blue Screen it was clearly mentioned that this happened due to Hantek Drivers (I am using the latest drivers from the Hantek websites).
I searched again and found that various other people has faced the same issue, and the suggestion was to use the OpenHantek6022 Software/Application.
So I started to make it happen.
And on the GitHub page they have mentioned the following line.
Since for Option-1 the document was available I choose Option-1 and referred the below pdf from the GitHub page
https://github.com/OpenHantek/Op ... 022_zadig_Win10.pdf
I used this Zadig tool to update my drivers, but my switch was still on position P i.e. un-pressed so what exactly happened that my Logic Analyzer drivers were updated, and now I can use OpenHantek Application in position P i.e. Unpressed but not the Logic Analyzer.
How I found this that this is the issue? when I tried to uninstall the drivers and deleted them from device manager, and I was randomly checking the USB VID and PID in both cases (it should display unknown device after deleting/uninstallation but it was Hantel6022BL with yellow exclamation mark for driver installation)
In normal case it should be like this.
Pressed: 04b4:602a Cypress Semiconductor Corp. (used for oscilloscope mode)
Not pressed: 0925:3881 Lakeview Research Saleae Logic (Saleae Logic VID/PID, so fx2lafw works out of the box)
But in my case for Not Pressed case the situation is that my ID 04b5, so I understood the problem that I created the problem myself using the Zadig tool accidently in a wrong way.
And then I started searching on internet how to revert this, because even after deleting and uninstalling the drivers the problem was not fixed.
Then I came across these two posts.
http://mirobetm.blogspot.com/2014/08/removing-libwdi-drivers.html
https://github.com/pbatard/libwdi/issues/134
There is some method given here, but before following that the problem got fixed.
The developer has asked to uninstall and delete several times, restart PC, and it should work.
Every time I am uninstalling/deleting the drivers from device manager, and re-plugging the device to USB, I am getting Hantek 6022BL but at last I get Unknown Device, and then I checked the VID of the USB Device and it was fixed back to 0925.
And then I manually installed Salae Logic Analyzer drivers and everything works fine.
When Button is Unpressed i.e. P position I installed Salae Logic Analyzer drivers
and when Pressed I installed the drivers given in Option-2 also available inside OpenHantek folder (not the Zidag method)
And now, everything is working fine.
OpenHantek
Salae Logic Analyzer
Again I am not really how I fixed this problem, but the problem was that I accidently changed the Logic Analyzer driver using Zidag tool. This I also cross checked in Events tab under device properties of device manager.
|