Electronic Engineer Discuss

Author: WisdomAugust
Print Prev. thread Next thread

Digital Multimeter HDM3055 Series Manual

[Copy link]

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
61#
 Author| Post time 2023-7-3 08:30:21 | Show all posts
Step 3: Press Probe and choose the probe type.
Additional settings for each probe type are described in the sections below.
For the RTD 2w or RTD 4w Probe Type...
If you choose to use an RTD, the menu will have a softkey to specify the RTD's resistance at 0
degrees Celsius (R0 ). Default 100 Ω


For 2-wire measurements, an Auto Zero softkey is available.
Auto Zero: Autozero provides the most accurate measurements, but requires additional time to
perform the zero measurement. With autozero enabled (On), the DMM internally measures the  
offset following each measurement. It then subtracts that measurement from the preceding reading.
This prevents offset voltages present on the DMM’s input circuitry from affecting measurement
accuracy. With autozero disabled (Off), the DMM measures the offset once and subtracts the
offset from all subsequent measurements. The DMM takes a new offset measurement each
time you change the function, range, or integration time. ( There is no autozero setting for 4-wire
measurements.)
For the Thermis2w or Thermis4w Probe Type...
For the Thermis2w probe type, an additional Auto Zero setting is available.


This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
62#
 Author| Post time 2023-7-10 08:19:32 | Show all posts
Step 4: Press Aperture and choose the number of power-line cycles (PLCs) to use for the
measurement. Only 1, 10, and 100 PLC provide normal mode (line frequency noise) rejection.
Selecting 100 PLC provides the best noise rejection and resolution, but the slowest measurements:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
63#
 Author| Post time 2023-7-11 08:16:48 | Show all posts
Step 5: Use the Units softkey to display temperature in degrees Celsius, degrees Fahrenheit, or Kelvin.
Thermistor Requirements
The DMM converts the measured thermistor resistance to temperature using the Steinhart-Hart thermistor equation:
1?T = A + B (Ln(R)) + C (Ln(R))3
A, B, and C are constants provided by the thermistor manufacturer and derived from three
temperature test points.
R = Thermistor resistance in Ω.
T = Temperature in degrees K.
Important: Use only a 5 k? 44007-type thermistor. This type thermistor has constants of A = 1.285e-3, B = 2.362e-4,
C = 9.285e-8. Using an incorrect type of thermistor can result in errors greater than 20 °C for a temperature being
measured of 100 °C.


58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
64#
 Author| Post time 2023-7-12 08:12:25 | Show all posts
Capacitance
This section describes how to configure capacitance measurements from the front panel.
Step 1: Configure the test leads as shown.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
65#
 Author| Post time 2023-7-13 08:29:39 | Show all posts
Step 2: Press



on the front panel.
Step 3: To null–out the test lead capacitance:
l Disconnect the + and - test leads probe end from the test circuit, and leave open.
l Press Null. The DMM will now subtract this null value from capacitance measurements.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
66#
 Author| Post time 2023-7-14 08:21:30 | Show all posts
Step 4: Press Range to select a range for the measurement. You can also use the [+], [-], and
[Range] keys on the front panel to select the range. (Auto (autorange) automatically selects the
range for the measurement based on the input. Autoranging is convenient, but it results in slower
measurements than using a manual range. Autoranging goes down a range at less than 10% of
range and up a range at greater than 120% of range. For capacitance measurements only, when
autorange is off, the instrument does not report an overload for readings greater than 120% of range.
Overload only occurs when the algorithm times out because the applied capacitance is too large for
the algorithm to measure. If you apply a DC voltage or a short to the input terminals in capacitance
measurement mode, the instrument reports an overload.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
67#
 Author| Post time 2023-7-15 08:14:25 | Show all posts
Continuity
This section describes how to configure continuity tests from the front panel.
Step 1: Configure the test leads as shown.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
68#
 Author| Post time 2023-7-18 09:41:15 | Show all posts
Step 2: Press [Cont] on the front panel to open a menu where you can enable or disable the beeper
for all functions that use the beeper (limits, probe hold, diode, continuity, and errors).



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
69#
 Author| Post time 2023-7-19 08:19:12 | Show all posts
Continuity measurements behave as follows:



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x

58

Threads

1280

Posts

1360

Credits

金牌会员

Rank: 6Rank: 6

Credits
1360
70#
 Author| Post time 2023-7-20 08:22:17 | Show all posts
Diode
This section describes how to configure diode tests from the front panel. The range and resolution
are fixed; the range is 10 VDC (with a 1 mA current source output).
Step 1: Configure the test leads as shown.



This post contains more resources

You have to Login for download or view attachment(s). No Account? Register

x
You have to log in before you can reply Login | Register

Points Rules

Dark room|Mobile|Archiver|Electronic Engineer Discuss

2024-6-6 01:39 GMT+8 , Processed in 0.179162 second(s), 22 queries .

Powered by Discuz! X3.2

© 2001-2013 Comsenz Inc.

Quick Reply To Top Return to the list