Title: Hantek DSO5202BM: Confusing vertical shift waveform. [Print this page]

Author: Oleg10011001    Time: 2017-1-10 14:23
Title: Hantek DSO5202BM: Confusing vertical shift waveform.
There is a strange shift in the vertical waveforms obtained from the second channel when the oscilloscope is set at more than 2V per cell. For example, look at the waveform in the following screenshots. In all of these cases, both the probe was connected to a generator, oscilloscope calibration, and both channels markers moved strictly in the center of the screen by pressing the corresponding "Vertical position".










Author: amy    Time: 2017-1-17 15:22
Please disconnect any probes or cables to the channels and press the UTILITY button, select Self Calibration option and follow the directions on the screen.

Author: Oleg10011001    Time: 2017-1-17 22:13
Thanks, this helped!