Title: Hantek6022BE Service call [Print this page]

Author: Chandrakant    Time: 2017-11-13 11:20
Title: Hantek6022BE Service call

Hello


This is  chandrakant from Bangalore, India.

Recently i have purchased Hantek6022BE (Serial No. 681057079) about one month back. Now it is not showing signal more than 5V (Saturating at 5V Max. )


Request you to provide technical support to resolve this issue on priority.


Thanks & Regrads,

Chandrakant




Author: amy    Time: 2017-11-13 13:09
Edited by amy at 2017-11-13 13:11

The maximum measurement vlotage of Hantek6022BE is +/-5V (without external attenuation). If the input voltage is greater than +/-5V, greater than 5V signal will be cut off.
You can use the probe with 10× attenuation to test.