Customer Consultation:

We have bought a CC65 current clamp, this unit have a 4% error at 50A, we have try to calibrate with or calibration shunt of 0.1% error adjusting the VR2 potentiometer, we can calibrate the output at 50A but it seems not linear and then it have error at 20-25A

This clamp has a lot of potentiometers inside, but Do you have a calibration manual for this clamp?



Author: ying    Time: 2022-6-15 21:48
Reply:

Please refer to the accuracy specification of the unit.

Accuracy:
         DC Current: 1m/10mA
±(1.5%±5mA) 10mA ~20A
         DC Current: 1mV/100mA
±(2%±20mA) 100mA ~ 40A
±(4%±0.3A) 40A ~ 65A
AC Current: 1mV/10mA
±(2%±30mA) 100mA ~ 10A (40Hz ~ 2KHz)
±(4%±30mA) 100mA ~ 10A (2KHz ~ 10KHz)
±(6%±30mA) 100mA ~ 10A (10KHz ~ 20KHz)
±(8%±30mA) 10A ~ 15A (40Hz ~ 20KHz)
AC Current: 1mV/100mA
±(2%±30mA) 100mA ~ 40A (40Hz ~ 1KHz)
±(4%±30mA) 100mA ~ 40A (1KHz ~ 2KHz)
±(6%±30mA) 100mA ~ 40A (3KHz ~ 5KHz)
±(8%±0.3A) 40A ~ 65A (40Hz ~ 20KHz)